scanner error No. 06332-00005

Recommended Posts

Noritsu S-1800 Scanner
I used "HIGH" resolution to scan 35mm film with AFC-II , and appear error No. 06332-00005 Light Source adjustment error. ( Zoom value abnormal) 
The images are out of focus. 
But I use "MID" or "LOW "resolution to scan films are fine. The Noritsu tech say [scanner unit] have problem.
Any suggestion ? 

Thank you.

Share this post


Link to post
Share on other sites

Hello. Can be scanner unit problem. 06332-00005 means light Source adjustment error. - the number of CCD loaded pixels increases and overs the standard value, because the scanner lens has problems and they affect the zoom value, or number of CCD loaded pixels decreases due to dust, incorrect attachment of AFCs, or problematic light axis.

Enter service mode. Then go maintenance menu. There go to scanner sensitivity check and test CCD and LED there. To see whole CCD area can remove film carrier.

Share this post


Link to post
Share on other sites

Thank you~Minilab Service

The Noritsu tech bring a HS-1800 scanner to me to test what problem.He put my S-1800 PCB board to his scanner. It work fine. So the problem is [scanner unit].

He say the [scanner unit] cost about USD$ 7000.

Can only change some part?not all the scanner unit?

Thank you

20200605-1800.jpg

Share this post


Link to post
Share on other sites

Scanner has many boards - scanner control pcb, scanner driver pcb, scanner main pcb and power suppliers. Do not know what your Noritsu technician tested.
When you had other scanner you could test most important parts - scanner control pcb, scanner main pcb ( J391249 ) and LED unit. Your B LED works at quite high value ( 233, when maximal is 255 ) . Before experiments need make both scanner backups. When replaced main control pcb need to load backup to restore scanner settings.

Share this post


Link to post
Share on other sites

The Noritsu tech just  put my [scanner driver PCB] and [scanner control PCB] to his scanner."HIGH" resolution is no problem.

The [Check Point] on 1800 scanner have 5 parts.So he say the problem is [scanner unit].

12103286470246.jpg

Share this post


Link to post
Share on other sites

No. 06305 -00002  Scanner Focus operation error.

I  got the new error message.

When normal open the scanner , It have about 3-5 seconds sound like zoom in and zoom out.

But it just have 1 second sound, and appear the error message 06305-00002

What is problem with this scanner?

Thank you

 

Share this post


Link to post
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
Sign in to follow this  

×
×
  • Create New...