fotomaria Posted August 19, 2011 Report Share Posted August 19, 2011 Hello All! I have a problem with my Frontier 340. It gives me this error "solution remains in tank" this hapens after he empties the catridges, after this error, he will let me print about 100-200 photos and the will say "number of prints excedes limit". In order to fix this i'm forced to poor water in the P1 tank (where the cartridge goes), because in the "Procesor Input Check" is showing "level low" for P1R. I changed the pump for the PSR because it was leaking an i thougt that was the problem. But it still gives me that error. Any sugestions? Thanks! Quote Link to comment Share on other sites More sharing options...
DTR Tech Posted August 22, 2011 Report Share Posted August 22, 2011 I would check over the rubber hoses to make sure none of them are getting kinked anywhere. The water supply hose can get kinked when you push in the tray with the filters, bellows and pumps. Sounds like it isn't getting water pumped into the tanks. It could also be a clogged hose somewhere or a solenoid valve in the block that is faulty. Kinks are most common. Quote Link to comment Share on other sites More sharing options...
ghn5ue Posted August 22, 2011 Report Share Posted August 22, 2011 Can't this also be caused by uneven levels in the mixed replenisher tanks? This happens because the pump outputs are not set correctly, and the machine is actually replenishing a different amount than it thinks. Next time the message occurs, drain the remaining replenisher (taps down below the cartridge). Then put in the new cartridge and mix. Once done you need to go into processor setup and run the pump output check. Measure the output in a graduate and enter the values and you should be set. Quote Link to comment Share on other sites More sharing options...
fotomaria Posted August 22, 2011 Author Report Share Posted August 22, 2011 Thanks for the answers. I forgot to mention, I checked the pumps and the output is the same as set, i haven't checked the water pump output, I'll do that and after I will check the hoses and all the other. Thanks! Quote Link to comment Share on other sites More sharing options...
sameer bahho Posted September 23, 2011 Report Share Posted September 23, 2011 THA SAME PROBLEM HAPPEND WITH ME I HAVE FRONTIER550 BUT THA MASSEG IS PROBLEM IN STIRING VALVE CHECK IT Quote Link to comment Share on other sites More sharing options...
Christina Rose Posted September 25, 2011 Report Share Posted September 25, 2011 drain out all the old chemical from Rep tank and make sure all the sensors are shows ON at lower side ( no need to drain PSR). Mix fresh chemical. before mixing check out the presents of any fungal formation inside the P2RA Rep tank. some times fungal formation given a wrong level sensing information and make similar problem. C..R Quote Link to comment Share on other sites More sharing options...
sisir Posted September 30, 2011 Report Share Posted September 30, 2011 My suggestion is also same as ROSE Quote Link to comment Share on other sites More sharing options...
sisir Posted September 30, 2011 Report Share Posted September 30, 2011 My suggestion is also same as ROSE Quote Link to comment Share on other sites More sharing options...
Bri Posted October 21, 2011 Report Share Posted October 21, 2011 I have had a similar issue myself. I agree with rose as well. Drain all replenisher ( NOT WORKING TANK) set a new cube and mix new replinisher. You have already confirmed that your pump outs are correct so now you need to varify that your rep rates are correct. They are set at such a low rate anyway that when they arent set properly one may run out before the other. Also don't make the machine wait if it ask for water PSR while mixing. Quote Link to comment Share on other sites More sharing options...
Geoffrey Posted July 22, 2019 Report Share Posted July 22, 2019 Hello I need help I have a Fuji 340 and its been leaking chemical P1 and I seem not to know the source of this leak and its becoming impossible to print photos with P1 leaking especially after a big number of print's. What could be the source of this leak? Quote Link to comment Share on other sites More sharing options...