ahmediraq Posted June 12, 2011 Report Share Posted June 12, 2011 [face=Times]when Im printting a picture some photo have a yellow colour im making a upkeep to corecet this error and it corected but at the next day the upkeep upnormal blue and im corectted it and my pictures its ok that error not constant help me pleas im from iraq thanks [/face] Quote Link to comment Share on other sites More sharing options...
yustas Posted June 12, 2011 Report Share Posted June 12, 2011 AOM driver is getting dead. The last call. Quote Link to comment Share on other sites More sharing options...
ahmediraq Posted June 13, 2011 Author Report Share Posted June 13, 2011 thanks but my problem at sum time and the temprtuer calebration for R,G,B, normal but my space tempretuer about 35 degree centegrate may be its problem help me Quote Link to comment Share on other sites More sharing options...
yustas Posted June 13, 2011 Report Share Posted June 13, 2011 if you look at working condition of laser you will see that there is a certian requirements exist: temperature - 18 ~ 30 degrees and humidity 70% at maximum. If you have ambient temperature higher than factory recommends - you may have problem of course. but if you decide now to chill out the lab - it will not help. So tyr to swap AOM driver first and find out which one is defective. how to swap I described here: http://www.minilablaser.com/problem_loc.htm Quote Link to comment Share on other sites More sharing options...
DTR Tech Posted June 13, 2011 Report Share Posted June 13, 2011 I agree with yustas. Quote Link to comment Share on other sites More sharing options...
ahmediraq Posted June 14, 2011 Author Report Share Posted June 14, 2011 thanks evry body the lesser history temp. Quote Link to comment Share on other sites More sharing options...
Kevbo37 Posted June 14, 2011 Report Share Posted June 14, 2011 It is definitely the AOM driver. These units are very problematic. Since the Fuji only has one AOM driver it is necessary to switch the inputs G for B and B for G. Then switch the outputs G for B and B for G. Your problem will now be a magenta/green shift. You must either replace the AOM driver or repair the unit you have. If you would like to seek a repair, please contact National Laser Co. Quote Link to comment Share on other sites More sharing options...
leandrocmalves Posted June 15, 2011 Report Share Posted June 15, 2011 Hi... Please verify paper white and paper black (with light), i change first P1 and P2, if problem back again, change AOm, and last change laser. LEandro Alves Quote Link to comment Share on other sites More sharing options...
ahmediraq Posted June 15, 2011 Author Report Share Posted June 15, 2011 THANKS ALL FRIENDS I SOLVED MY PROBLEM BY COOLING THE LESSER UNIT TO TEMP. (25) AND MY PRINTING NOW VERY NORMAL CONSTANT COLOR THANKS THNKS Quote Link to comment Share on other sites More sharing options...
yustas Posted June 15, 2011 Report Share Posted June 15, 2011 well.... may be begin to produce cooling units for minilabs... real free niche in market... Quote Link to comment Share on other sites More sharing options...
ahmediraq Posted June 17, 2011 Author Report Share Posted June 17, 2011 [face=Times]ok but i replace the OAM drive and the problem solved and my print it is ok but how I test the OAM drive really and how I Protect it . help me pleas ...... [/face] Quote Link to comment Share on other sites More sharing options...
Ivcha92 Posted July 23, 2011 Report Share Posted July 23, 2011 I have found fast solution if you need to finish your work fast. Just swap BNC connectors on left and right sides (upper b to upper g and lower b to lower g)on AOM driver and print some condition tests. You will be able to finish work an maybe work for a couple of days but AOM driver still need to be replaced. Quote Link to comment Share on other sites More sharing options...