mike11918 Posted February 17, 2012 Report Share Posted February 17, 2012 I've been receiving error "Solution level of the processing tank goes below the processable level. P1" I've cleaned and checked the floats in the P1 processing tank and it's fine. I'm not sure what the problem could be. Any ideas? Thanks. Quote Link to comment Share on other sites More sharing options...
Magenta Posted February 18, 2012 Report Share Posted February 18, 2012 solution level is low ;D If the level is low and the float is not stuck, check the rep outputs and water refill system. As you use the machine P1 will drop and must be replenished, Quote Link to comment Share on other sites More sharing options...
mike11918 Posted February 18, 2012 Author Report Share Posted February 18, 2012 I have checked. All outputs are correctly replenishing as they should. I even tried increasing the P1 output, but that didn't result in anything positive either. Error still shows. Quote Link to comment Share on other sites More sharing options...
Magenta Posted February 18, 2012 Report Share Posted February 18, 2012 Remove the float switch and check operation in input check, the floats should change state when moved, also try the float-switch in some water, they sometimes don't float, Quote Link to comment Share on other sites More sharing options...
mike11918 Posted February 18, 2012 Author Report Share Posted February 18, 2012 Floats do change state when moved in input check. Float switch floats ok. Change is only detected when I force float down low or high up. Quote Link to comment Share on other sites More sharing options...