Ikaianer Posted June 25, 2011 Report Posted June 25, 2011 Good morning, dear Friends your help is needed. I am running a QSS 3011 in central america on 220V and 60Hz. Since power is very unstable here I have bought a UPS unit with 10KVA and 8KW. Now this unit is producing a 220V 50 Hz current. I learned that the machine can handle that by changing the gear of the drive motor. OK I have opened the dryer section spotted the motor and the gear. But If I turn the gear onto the slower speed side it would not fit back with the other corresponding gear. So what is the trick??? Can somebody please help me? Thanks in advance and hope you had a great fatherday. ) Andy
Dave S Posted June 25, 2011 Report Posted June 25, 2011 The drive motor fixing bracket is slotted, simply loosen the 4 screws and lower the drive motor, ensure there is a little bit of backlash between the gears.
yustas Posted June 25, 2011 Report Posted June 25, 2011 strange... I thought it is just enough to reinstall and turn the double faced gear from 60Hz to 50 Hz position and all will work fine... there is a scheme sticked on the panel...
Ikaianer Posted June 26, 2011 Author Report Posted June 26, 2011 Hi guys, thanks for the help the unit with 10KVA UPS with 50Hz is running just great ) Hope I can help you one day. Wish you a great day. Andreas