Big Dave Posted May 3, 2011 Report Posted May 3, 2011 OK Guys, Bad News & Good News & a question. Bad News: I've got a dreaded dead section on our MLVA head - and it is affecting the uniformity correction, and gives us "6088-0006 Uniformity Correction errors" every morning - but we can live with it if we have to. Good news: It only affects a small 1.75mm to 2.25mm END section on 6" papers coming out of one lane of the machine. My question, can I "move" or shift the paper/advance/imaging of our 6" paper in ONE (or both) of the lanes, such that the affected area is not in the damaged imaging area of the MLVA head ? If you look at my attached image you can get a better idea of the situation.... It seems like it could be a very easy fix, but I don't know what to adjust to do this. I just need to move it 3mm to the left. Anyone know how to accomplish this ? And, anyone have a working MLVA head (for a 2901) they're willing to part with for a reasonable price ? Thanks guys !
aussiecameras Posted May 4, 2011 Report Posted May 4, 2011 Hey Big Dave, that doesn't look like a standard MLVA problem to me. I have personally looked at two 2901's with a dead area of MLVA and each time it has been 8mm wide and the area has been lighter than the other areas. And on this forum and the other one many more it has always been 8mm. I think that the 8mm gap is how the array system works so I would suspect some thing else than a dead MLVA. One thing I remember looking at with my problem was looking at the actual rows of LED's coming out of the MLVA unit. I used a mirror to look at the light coming out of the MLVA unit and from memory there were three rows of LED's and at the spot the problem was only two of the rows were working. But I'm just a lab owner, so over to all the experts! Whatever the problem, if you can move it all over a bit everything would be fine! I reckon it is something else though, not the usual 8mm MLVA line of death ! And it is Death! Good luck !
Big Dave Posted May 5, 2011 Author Report Posted May 5, 2011 Hi AussieCameras, Thanks for the tip to look at the MLVA with the mirror. Tried to check this morning, but when we do a "Pressure Guide Cleaning" to look at the MLVA head, all the lights are off. What did you do to check the MLVA lights ? The pressure guide is in the way and have no access to view the MLVA head. Am hoping it's not dead, and just some other problem. We kept getting Uniformity cailbration errors every morning since this mystery spot showed up, and may be fortunate that it's happened at one end of the MLVA head that is affecting only a small portion of the printed image. Here are the error messages we got just before the spot started: 6088-0006 - Uniformity Correction error 6315-0084 - Setup Filter dispersion correction error (Positive) 6900-0001 - Main control system error 6310-0084 - Scanner area registration error 6322-0008 - Scanner input balance error I'm hoping for some suggestions on shifting the image, and how to work around those failed uniformity calibrations as well as, get a lead on a replacement MLVA head or a better bead on fixing these problems. If anyone thinks it's the Image Processing PCB or other possible PCB, let me know.
srtsr Posted May 10, 2011 Report Posted May 10, 2011 Do these lines show up on the same spot on a 12 inch wide piece of paper? Make sure reflector in MLVA lamphouse is very clean as well as dichroics. Lamp must also me in good condition. Make sure there is no dirt on top of filter wheel glass. Lamp should have no more than 900 hours on it. Make sure to clean everything real well in MLVA lamphouse not scanner lamphouse, have new bulb and then call noritsu to have them give you password of the day and have them walk you through drive voltage registration and uniformity.
Dave S Posted May 10, 2011 Report Posted May 10, 2011 You have several issues going on here! Did all the errors start at the same time? Have you tried loading in your backup data? 6900-0001 main control system error it says ARCNET or LVDS cannot be used. Check the drivers for the LVDS and ARCNET cards are loaded properly in Windows. There is a long list of boards the manual says could be failed. D-ICE PCB Printer control PCB PCI ARCNET & LVDS cards Image processing PCB. The other errors relate to the film scanner, mostly pointing towards a lack of communication between the scanner and the rest of the machine. D-ICE PCB Image processing PCB Check all the connections around the image processing PCB. Remove and re-fit the D-ICE PCB in case of poor contact. With regards to moving the position of the print, that’s easier said than done! You have very little correction to work with around 1mm, and it will likely start giving you white lines on the opposite track. I personally wouldn’t go down that road. In most cases the MLVA head can be revived by doing the uniformity pattern 2, 3, 4 & 1. First thing to try is fit a new MLVA lamp, clean the junction between the underside of the filter wheel and the fibre optic cable, this can get very dusty greatly reducing the light going to the MLVA head, make sure the dichroic filters, reflector, and the filter wheel glass window are clean. After cleaning do the MLVA lamp change setup. See if it now completes the uniformity or not. The uniformity correction error can also be caused by a problem with the image processing PCB, so ideally you need to sort out all the other errors before attempting to do anything with the uniformity problem.
Big Dave Posted May 10, 2011 Author Report Posted May 10, 2011 Hi guys, Thanks for the responses ! We've been busy with Mothers Day photos. Quick status report : we replaced the scanner lamp and did lamp change and most messages have "gone away". Only message we get now is at the beginning of morning start up (past 3 days in a row) is : 6088-0006 - Uniformity Correction error Then after successful daily setup it always wants to do a Uniformity Calibration - usually finishes in about 10-15 minutes with a Uniformity Calibration error. I would love to perform Uniformity Pattern 1, but without a flatbed scanner, we cannot do it. We'll try everything else first and let you know how it goes. For now, we've been cutting off 2.5mm off the 4x6's and none of our customers have complained... yet. After we finish our 6" papers, we're going to change to 4" papers if we can't get this figured out.
Dave S Posted May 10, 2011 Report Posted May 10, 2011 You can do uniformity pattern 2, 3 & 4 and 1 without the flatbed scanner, assuming the calibrator does a good enough job with the uniformity. I'll dig out the info for you if you're interested.
Big Dave Posted May 11, 2011 Author Report Posted May 11, 2011 Dave S, It would be much appreciated if you could find those documents. Some questions I have: With the MLVA head in it's current state, how much more life is left on our MLVA head ? What are the chances of another section going bad ? Thank you.
Big Dave Posted May 13, 2011 Author Report Posted May 13, 2011 Hi Dave S, Did Load Voltage Adjustment - OK Uniformity pattern 2 - OK Uniformity pattern 3 - ran through the cycles 3 or 4 times, then got this: (see photo) It crashed the QSS_SEQ.exe.... It's late, so maybe I'll try again tomorrow night.
Dave S Posted May 13, 2011 Report Posted May 13, 2011 Try following this, the problem may have happened because the MLVA data was not initialized first. Obviously make a full service backup before you do anything, just in case of any problems. Uniformity pattern 2, 3 and 4 IMPORTANT • The uniformity patterns 2, 3 and 4 are always to be used together. Be sure to follow all the steps below. 1. Initialize the MLVA data. Manual 3066 Menu,Extension,Setup, Data Initialization, MLVA 2. Delete unf_m_nai.dat in C:\NORITSUKOKI\QSS29\Data. IMPORTANT • Be sure to delete unf_m_nai.dat. The data may not be updated and the print quality can not be brought back to the reference level if unf_m_nai.dat is not deleted. 3. Perform inner calibration correction adjustment. Manual 3220 Menu,Extension, Maintenance, Various Adjustment, Printer Mechanical Adjustment 4. Carry out the Load Voltage Adjustment. Manual 3062 Menu, Extension, Setup, MLVA setup, Paper Specification Registration/Setup, F-Functions 5. Carry out the uniformity pattern 2. NOTE • Follow the instruction on the display. • It will takes 15 minutes at most. 6. Carry out the uniformity pattern 3. NOTE • Follow the instruction on the display. • It will takes 60 minutes at most. 7. Carry out the uniformity pattern 4. NOTE • Follow the instruction on the display. • It will takes 30 minutes at most. 8. Carry out the uniformity pattern 1. Creating the FB coefficient file The uniformity pattern 1 (flatbed scanner) is omitted, and the time can be shortened when uniformity patterns 2, 3 and 4 are done and it is judged that stripes and unevenness disappeared. Create the FB coefficient file according to the following procedure. 1. Delete UNF_FB.dat in C:\NORITSU KOKI\QSS29\DATA, or change the file name. 2. Copy the UNF01.dat in the same folder and change the file name to UNF_FB.dat. IMPORTANT • When using another folder for the copy, change the file name to UNF_FB.dat and save it to C:\NORITSU KOKI\QSS29\DATA. And, when you change the file name, be careful not to mistake the file name. If the folder to save is mistaken or the file name is wrong, when performing the Uniformity Pattern 1 (Flatbed Scanner), an error can occur and cannot perform. 3. Carry out the Uniformity Pattern 1. (1) Select the paper magazine. (2) Press the Skip button to skip the correction by the flatbed scanner. The calibrator measures the state. (3) Register by YES. 4. Check the print. shutter 1
Big Dave Posted June 7, 2011 Author Report Posted June 7, 2011 Status Update: Better, but no Fix of original problem. Hi Dave S, Yesterday morning we had an additional section on the left lane that had a bad section so all prints were bad. We cleaned the "Integrator Section" and we ran through your instructions above, but it appears that the Uniformity Pattern 3 & 4 did not complete 100% successfully like Uniformity Pattern 2. ie: After completion of Pattern 2, it would say if it was successful or not and wait for you to hit "YES" to go back to the Uniformity menu, but on Pattern 3 & 4, I'm assuming it completed (didn't lock up the machine) but it went straight back to the Uniformity Menu without letting me know if it was successful or not, or asking for me to hit "YES". Also, the file: unf_m_nai.dat - was deleted as instructed, but it was never re-created by any of the Uniformity Patterns 2, 3 or 4. (I checked this morning after daily setup, and it still was not there.) and the file: UNF01.dat - was dated 4/29/2011, and a new one also was not re-created. Uniformity Pattern 1 gave me problems and I was not able to complete it without the QSS wanting to print / use the flatbed scanner, so I gave up on it. BUT the process did "fix" the new problem on the 2nd Lane so we are back in business, but the original problem section on the far right of the 1st Lane still has the issue. I'll try this again when I have more time.... We just got some new HP kiosks and I'm spending way too much time getting that system setup, but looking forward to the advanced capabilities it has over our current CT-SL/CTX kiosk. Thanks much for the above instructions.
Dave S Posted June 8, 2011 Report Posted June 8, 2011 I'm wondering if the calibrator is not moving right to the end of the MLVA head, maybe it's getting stuck etc. You can test it in output check mode you'll need to put a magnet on the sensor to run it with the covers open. It should move smoothly without any kind of jerkiness. Maybe a dab of oil on the calibrator worm shaft would help. Not sure how easy it is to get to the worm shaft without lifting up the whole MLVA head. It's strange that the unf_m_nai.dat file was not re-created, maybe it will only re-create this file after doing the uniformity pattern 1. When doing the uniformaty pattern 1 you do actually have to make the test prints, but you skip the step when it asks to measue with the flatbed scanner. What this is actually doing is recording the current state of the uniformity as the reference file. So it aims to bring it back to this state each time it does the uniformity.
Big Dave Posted June 8, 2011 Author Report Posted June 8, 2011 When doing the uniformaty pattern 1 you do actually have to make the test prints, but you skip the step when it asks to measue with the flatbed scanner. Yes, the test prints came out, but then it asks to measure on the flatbed scanner, it only offered "YES" or "NO to cancel" I kept hitting "NO" at this point, confirmed the "are you sure you want to cancel" and it would exit the Uniformity pattern 1 process completely. Next time, I'll hit "YES" at that point and see what happens. The oil for the calibrator... I've been using a "white teflon" grease for the processor bottom rack gears on the main drive shaft (a noritsu tech at PMA said it was ok). Would this same oil work, or would you recommend something else for that ?
Dave S Posted June 8, 2011 Report Posted June 8, 2011 I agree it's not very clear what to do when carrying out the uniformity pattern 1 without a flatbed scanner, it always confuses me as well! It's a while since I last did it so I can't remember the exact steps. Next time I'll write it down! No I wouldn't use the white grease it is far to thick. Just use a very small amount of thin oil to lubricate it, say 1 drop every inch. Japan use a special grease that is something like $200 for a 1oz tub! You normally get a small pot of oil with the film processors, that is ok to use. Be careful with the white grease, Noritsu supply a small tube of white grease normally marked 'ball grease' but this is not for the rack drive gears, it is for lubricating the magazine tray runners and the arm unit runners etc. Don't use this grease on any plastic parts or get it on the tank as it causes the plastic to go brittle and then it can crack.