It is most likely a wiring problem with the red laser. Check the two small white wires coming out of the laser unit and also the Red and Black pair . These are the sensor and heater wires, respectively. Also check the two small black wires (attached together), they are the unit thermo sensor.

Link to comment
Share on other sites

Yes all correct...

Also:

Open the top cover (not a cover of a laser unit!) and get access to the laser area.

Inspect connectors 133 (TEC) and 134 (termistor) at the Laser I/O PCB #J390641.

Reconnect them for in case.

If all these actions will not help - Laser I/O PCB #J390641 is probably defective.

Link to comment
Share on other sites

Thank you Kevbo & Yustas, I checked all the cables you talked about. I swapped the laser units,changed the Laser I/O PCB & I got the following messages: 6075-0004 G Laser control error

                6074-0004 B Laser control error

                6073-0001 Synchronous sensor error

I went on to put back the old laser together with Laser I/O PCB from another machine, again I get the errors:

             6073-0001 Synchronous Sensor error &

             6107-0001 R Laser temperature is out of range.

So please help me because the machine cant print right now

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...
DL650
DX100