photopeter Posted May 13, 2010 Report Posted May 13, 2010 Hello all, can anyone help me. CD tank on my V50 will not stay 37.8c that I have it set to. It fluctuates between 36.5 and 38.5, sometimes higher and gives a solution temperature error. Sometimes it will stay at the right temperature for 10-20 mins, then it just goes up and down and doesn't steady itself. I initally thought this was a circulation error so I have drained the tank, thoroughly cleaned it, blowing water through all the pipework, changed the chemical filter, I even took both circulation pumps off the machine, opened them up and made sure everything was clear. Put the machine back together with fresh chemistry and still the same problem occurs. I have checked the temperature with a digital thermometer and it matches what the V50 says, so I know the thermostat isn't at fault. I know the circulation pumps are working because if you manually turn them on and off in output checks menu, you can see the chemical moving in the compartment where the filter goes. What I don't understand is the speed it can go from 37.8 to 38.5, 5-10 seconds. But when it is warming up in the morning it takes a lot longer than that to cover 0.7c. Does anybody know what else I can check?? Sorry for such a long post, just wanted to give you as much detail as possible. Many thanks
Sait Terekli Posted May 13, 2010 Report Posted May 13, 2010 dear photopeter, did you open and clean circ. pump? if yes, you should change SSR relay of developer. Sait Terekli , istanbul mobil:+90 533 712 44 27
photopeter Posted May 13, 2010 Author Report Posted May 13, 2010 Hi, thank you for your reply. I have cleaned both circulation pumps and they work fine. I don't think it is the relay because it does stop heating. I thought if the relay was fault it would just keep heating and not stop. It's the speed it gets from 37.8 to 38.5 that worries me, the heating element isn't capable of heating at this speed which indicates a circulation or thermostat problem but I think I have checked everything. Could there be anything else? Thanks
Sait Terekli Posted May 13, 2010 Report Posted May 13, 2010 work without chemical filter, try it and look. if same problem , SSR is out of order, not work well. and look behind of board on SSR? sAİT
photopeter Posted May 13, 2010 Author Report Posted May 13, 2010 OK, i have another board, I'll try that. Thank you for your help, I'll let you know if it works. Thanks again
Dave S Posted May 13, 2010 Report Posted May 13, 2010 This is a caused by poor circulation. Check the sparge pipe that sits inside the CD tank on the right hand side is not blocked, it has lots of small holes that can easily get blocked, reducing the circulation.
photopeter Posted May 13, 2010 Author Report Posted May 13, 2010 Hi there, thanks for the reply I have checked that, when I take the pipe off the circ. pump and blow through it, I get little jets of liquid coming out of the sparge bar so this seems to be clear. Thanks
Dave S Posted May 13, 2010 Report Posted May 13, 2010 It's still worth actually removing the sparge pipe, and cleaning each hole with a paper clip etc, there are a lot of holes, so it may look ok when inside the machine, but 50% of the holes could be blocked. Other things you can check are the pipes going to the top of the pumps are not squashed flat at any point. Check the impellers inside both pumps spin freely on the shaft, there is often a build up of silver on the shaft causing the impeller to not spin freely, which will cause a reduction in the circulation.
photopeter Posted May 14, 2010 Author Report Posted May 14, 2010 Just to update, Took the board off and instantly saw the problem. SSR relay 1 has a black spot on one of the pins, looks like a fault has occurred. I have put a new board in and just waiting for the chemical to heat, hopefully not over heat !! Thank you to SAIT for solving the problem, my next question is: where can I buy a replacement relay to solder onto this board? Noritsu want £1000 for a new board and do not sell the seperate components. Thank you to all that helped.
Sait Terekli Posted May 14, 2010 Report Posted May 14, 2010 where are you from? i live in istanbul, Turkey. you can get here much more cheaper or you can get from me second hand board. mail is:saitterekli@hotmail.com if you have any question mail to me please. Sait
photopeter Posted May 14, 2010 Author Report Posted May 14, 2010 Thanks Sait I have located a new relay from an electronics company here in England. Thanks for your help
Dave S Posted May 14, 2010 Report Posted May 14, 2010 Normally when the SSR has totally gone the temperature will keep rising and rising until the safety stat cuts the power to the heater (normally around 40oC) Although I have had a faulty SSR that would only partially switch off, it was letting through 100V which most of the time did not cause a problem, I guess the power leak was about the same as what the machine needed to keep the tank at the correct temperature! The tell tale sign of a faulty SSR is the colour of the SSR goes slightly brownish. The CD heater is SSR1