I found a problem in NEW colorimeter of QSS 32, It read 23 steps instaed of 22 and hence reading error displayed. What could be the reason? why it read 23 steps?

Any expert?

Link to comment
Share on other sites

I am getting concerned that you have so many problems yet you have changed so much on your machine. Are you sure when you have changed such PCB's as Image control that you have loaded the system software and re flashed the various PCB's and reloaded any backups.

Looking at this fault in isolation, the colourimeter has a calibration sequence in the service mode to ensure that the test strip stops in the right place. It entails drawing a line on a piece of paper followed by a 2nd one and adjusting the pulses of the stepper motor so the line stops directly under the head.

In your case it sounds like it has no, or the wrong data