Calibration Error 6187


Videoteks

Recommended Posts

I have a 3202PRO I have received an error in the last couple of days that I am not familiar with and there is no information in the trouble shooting manual. The error is CALIBRATION ERROR 6187 it occurs when I do an emulsion change on 6" paper which is my set-up paper. Has anyone had this error before? Any help will be much appreciated.

Link to comment
Share on other sites

In the manual it says 6187 calculation error.

It says problematic setup calculation result detected.

Perform initial setup again.

Check your chemistry is all in control.

Clean the colorimeter calibration patch with a soft cloth and a little water.

Try loading in your backup data.

If none of the above helps, the initial setup (in the service mode) will have to be done again.

It could also be a problem with the colorimeter unit.

Link to comment