Posted

hello

i have qss3202.

the donge of soft dose not work it give me the message.capacity booster key was not detected.

what is the problem

any one can help me

thanks >:(

Posted

it's a optinnal software to make your 3202 faster, the dongle are key ( booster capacity ) instaled to the back of the computer in LTP port com and for instal it you need a software option

but the dongle key are not instaled in the manufacture.

bye frenchman

Posted

I think it is unlikely the dongle has anything to do with the main control system error.

If the dongle is faulty or the software for the dongle is not loaded you will get the capacity booster key was not detected error message.

What is the full error number for the main control system error?

Frenchman: I don

Posted

because noritsu as detected a problem with the new P1 chimistry kodak V 2.1 .when you change the tank capacity in total replenishment the follow alarm appear 6900 main control system error .but the system run normaly after close popup windows

the problem can be resolved by change the capacity tank 9999.9ml to 3999.9 ml (the memory capacity in processor control can't accept more that's 4000)

bye frenchman

     