Uniformity 2901(Uniformity Correction Error : 6088-0024)


John1972

Recommended Posts

Hi I have this problem with a 2901 team, when I do the uniformity throws the error (Error Correction Uniformity: 6088-0024) and when I adjust gains casts service mode (Auto Gain MLVA execution error: 6029-0003)  I try to make the uniformity Setup mode service also throws UNIFORMITY PATTERN 1 (Uniformity Correction error: 6088-0024).

 

Thanks in advance

 

 

 

 

 

Link to comment
Share on other sites

Clean MVLA head and try to follow these steps:

1. Initialize the MLVA data.
2. Del unf_m_nai.dat (on C:\Noritsukoki\QSS-29\Data folder)
3. Perform Inner Calibration Correction Adjustment.
4. Carry out the Load Voltage Adjustment.
5. Carry out Uniformity Pattern 2.
6. Carry out Uniformity Pattern 3.
7. Carry out Uniformity Pattern 4.
8. Del UNF_FB.dat (on C:\Noritsukoki\QSS-29\Data folder)
9. Copy the UNF01.dat file and paste it on desktop. 
10.Rename "UNF01.dat" on the desktop to "UNF_FB.dat"
11.Move the renamed "UNF_FB.dat" file to Data folder mentioned in item 8.
12.Perform the Uniformity pattern 1: 
- Choose the paper magazine. 
- Choose "Pass": to Cancel and skip the test print.
- Start the Uniformity measurement with the calibrator.
- Press "YES" key
13. Confirm the print quality.
Link to comment
Share on other sites

  • 2 weeks later...
  • 3 weeks later...

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...
DL650
DX100