Guest bigrath Posted July 24, 2014 Report Posted July 24, 2014 We just replaced the image process pcb board, the 35mm carrier and 35mm slide carrier work fine. The 120 carrier gives us this error message. We changed the scanner lamp and cleaned the carrier. Any idea? Thanks, Harry
Minilab service Posted July 25, 2014 Report Posted July 25, 2014 Can be something wrong with carrier or scanner alignments bad. 6310-0075 - scanner area registration error, the image is out of alignment. Put focus adjustment charts and look tilt - swing adjustments and light axis adjustment menu.
Guest bigrath Posted July 25, 2014 Report Posted July 25, 2014 I have never seen the focus charts where are they found?
Minilab service Posted July 25, 2014 Report Posted July 25, 2014 Hello. Engineers have adjustment charts ( 135, APS, 120, mounted slides ) . I have them too. Just now saw this fault appeared when you replaced image processing pcb. Which model you put there - old ( with RAM modules ) or new ( without them ) . If older model try clean RAM modules and try switch ( two the same RAM's ) . Do not try put there computer RAM or put these RAM modules to computer. If you do it your board will be damaged. Try go to service mode, software upgrade from HD and upgrade boards in image processing section. Also can switch image correction pcb ( have two the same there - one for scanner and other for digital ) . If all will be well, but still error might be image processing is faulty. Can test it on some other QSS29 minilab. Can't test on 2911, 2921 ( Kodak branded ) . Kodak minilabs work with Kodak scanners and instead image processing pcb have DLS boards.
Guest bigrath Posted July 30, 2014 Report Posted July 30, 2014 The image process pcb, d-ice pcb and 2 image correction boards were replaced. Everything works but the 120 carrier. The board has built on ram. We need the area registration done. Is there someone in the New York area that has the 120 focus chart?
kodak_service Posted July 30, 2014 Report Posted July 30, 2014 Worked 120AFC earlier at this minilab or it is the first installation?
kodak_service Posted August 1, 2014 Report Posted August 1, 2014 Try to read backup. Also check the position of 120AFC in scanner.
Guest bigrath Posted August 3, 2014 Report Posted August 3, 2014 Did read the backup, no change. What do you mean by check the position of the 120AFC in scanner
kodak_service Posted August 4, 2014 Report Posted August 4, 2014 If 120AFC isn't inserted against the stop, it can be a cause of error.