bshots1 Posted January 9, 2015 Report Posted January 9, 2015 I have a 3001 with sm chemical, what is the replenishment rate for CD-W and STB-W?? my print coming out blue and I know is not the laser problem, so I believe is my replenishment is not setting correctly (after 12 years owning the machine) CD- A - 3.01 CD- B - 5.50 CD- C - 5.81 CD- W - 50.27 STB - 1.49 STB - W - 192.25 THX FOR CHECKING!
Big Dave Posted January 9, 2015 Report Posted January 9, 2015 Check the kodak website I think this is the correct publication for you. https://www.kodak.com/global/en/professional/support/techPubs/cis190/cis190.pdf
bshots1 Posted January 10, 2015 Author Report Posted January 10, 2015 thank you thank you!! I will check on that.
Dave S Posted January 11, 2015 Report Posted January 11, 2015 I would be checking that the pumps are pumping out the correct amounts. Use a small measuring cylinder and take 3 readings add the 3 readings together and divide by 3 to give you the average. If it is under pumping, check that the white tap/filters are clean, then check that the poppet valves and seals are not worn out. See attached.
bshots1 Posted January 16, 2015 Author Report Posted January 16, 2015 great help sir, I took the CD-W out all day to measure the total amount of the day, we printed about 40-4x6 and 5-5x7, the machine pump out 4 liter of water to the tank!!! also shows: 5527 - STB2-W 5715 - STBW-W 5710 - CD-W 5713 - STB2-W all filters, screen, water hose and poppet valve has been replace/clean , production has no problem(since I redirect the water to a refill bottle),what else I can do to stop the alarm and stop the over filling??? thx!
Big Dave Posted January 16, 2015 Report Posted January 16, 2015 Just a comment, but you never mentioned that you were having an alarm all day long, or that it was over filling the working tank with water. If you are having a problem with the machine pumping excess amounts of water, maybe the float valve inside the working tank is stuck in the down position ? Check / clean that and let us know what happens.
Dave S Posted January 17, 2015 Report Posted January 17, 2015 Yes Big Dave is quite correct, it is the float switches in each sub tank that control the water re-filling. The other cause is if the level of the machine is way out, check it with a spirit level.