sapsi Posted October 20, 2016 Report Posted October 20, 2016 Hi!The racks not start, receive no error about transport board, only to calibrate chemicals. When i put the magnet, the racks start at a speed of 70-80 hertz for a few seconds, then stop.The chemicals is set to 20 sec developer, so the speed of the racks it should be 180 hertz. I started dksdiag, all is ok, changed the racks motor, same problem. Any help is appreciated. Thanks.
fotoart.hr Posted October 22, 2016 Report Posted October 22, 2016 (edited) I had the same problem. It was a problem in the power source. Arrived the lower voltage to the motor Edited October 22, 2016 by fotoart.hr
sapsi Posted October 24, 2016 Author Report Posted October 24, 2016 I connected separate power source with reglabile voltage to racks motor. The problem i think is "L4970A 10A SWITCHING REGULATOR" on PCB.