Hi guys, I have problem with Fuji Frontier 570. It is using replenishing chemistry incorrectly. 

 
I am getting error:
E-2663 P1R replenishment pump (PU721) or lower limit sensor (FS724) abnormal. About 510 3R-size sheets can still be printed without replenishment. Conslut your technical representative.
 
When I check level of replenishment chemistry in containers I can see the diference betwen levels, there is more P2RA and P2RB than P1R. The mos of chemistry is P2RB. 
I did pump calibration before I put new chatrige. I am using fuji hunt cp-49hvII]
 
Thanks
Posted

You can

 

Clean the level sensors

 

Run the PSR air filter - Front panel of the Fr570, there is a PSR air option.  Press menu 5 times and enter, then scroll down using the arrow keys.

 

Make sure there's no blockage in the hoses or the replenisher pumps (red Iwaki pumps)

Posted

If you pull off the back cover where the power switch is.  There is a number of float switches for the replenisher section also.  You can see what ones work correctly.  They're yellow and have white floats; they they're high, the processor output check for that replenisher would say YES.

Posted

I clean already, and I didn't fix the problem. In my opinion the problem is less water in P1, when machine cleaning the cartridge with chemistry. 