TheGothicThe Posted May 27, 2017 Report Posted May 27, 2017 Hi Wondering if someone could shed some light on Error Code E-4764 on SP-3000 Thanks in advance. Ahtong 1
Magenta Posted May 27, 2017 Report Posted May 27, 2017 From the manual E-4764 The external memory of GIA circuit board is abnormal. Restart the system. If problems persist, consult your technical representative . Result of GIA23 circuit board external memory R/W was NG. (Causes of the error message) 1. Poorly inserted GIA23 circuit board into slot of GMB23 circuit board 2. Faulty GIA23 circuit board 1. Insert the circuit board into the slot correctly. 2. Replace the circuit board. Ahtong 1
TheGothicThe Posted May 27, 2017 Author Report Posted May 27, 2017 Many thanks, Could this cause E-4748 as they both are linked somewhat together. Ahtong 1
Magenta Posted May 28, 2017 Report Posted May 28, 2017 E-4748 The AE retry-over occurred. Restart the system. If problems persist, consult your technical representative. At the time of AF, scanner correction, brightness correction, or AE (Causes of the error message) 1. Faulty harness between CCD23 and GIA23 circuit boards 2. LED light amount too large 3. Faulty CCD 1. Reconnect or replace the harness. 2. Perform Menu 0349 “LED Light Amount Adjustment”. 3. Replace the CCD unit. I suggest you reseat all cables and PCBs as a first step
TheGothicThe Posted June 1, 2017 Author Report Posted June 1, 2017 Thank you. I can confirm that it has been sorted with re-doing the cables and cleaning tower unit
Recommended Posts
Create an account or sign in to comment
You need to be a member in order to leave a comment
Create an account
Sign up for a new account in our community. It's easy!
Register a new accountSign in
Already have an account? Sign in here.
Sign In Now