vincent Posted September 21, 2017 Report Posted September 21, 2017 Dear, i search all post on the section but i can not find a solution. Sorry again, i create a new post. Hope you will understand. My 3202 was very slow for printing and sometimes stop priting since a few months.. Even the order have finish, print are not finish yet and machine stop printing. So I decide to change my RAM first, before i have 2Go, now i put 4Go. I move carefully the machine and take care the cable behind. First machine was ok, print were a little bit faster but not like before. After i put a second HD for making a backup, when it start, i have these 2 errors: first :6900-0001 Main control system error second: 6901-0008 Arcnet communication error. I turn off machine, but same problem. I unplug my second HD, and start again, error still here. I check all fiber cable behind machine and the scanner but still problem. I check the PCB arcnet in windows, it looks ok anyone have an idea please? Thank you
Minilab service Posted September 21, 2017 Report Posted September 21, 2017 Try to reinstall board driver - remove and install again. Then check connections using ARCNET test software. 6901-0008 means no ARCNET communication between printer control pcb and processor control pcb. Only if all OK when you run ARCNET test software you can try to run working software.
vincent Posted September 22, 2017 Author Report Posted September 22, 2017 thanks for your reply. I put the pcb which connect one plug and optical fiber to another PCI slot in the PC. I check all optical fiber ( behind PC, in the top of the scanner, behind under the monitor, behind magazine in printer, under dryer). I clean all the plug for the optical fiber. All the black optical fiber have light, the white don't have. Same problem When the soft start, the printer and the scanner don't make any noise. Usually they do thank you
Minilab service Posted September 22, 2017 Report Posted September 22, 2017 Did you read my previous message? There I wrote do not run Noritsu software ( remove QSS from Start, Program, Startup ) . Run ARCNET test software. There you should not have errors.
vincent Posted September 22, 2017 Author Report Posted September 22, 2017 sorry, where is located ARCNET test software
vincent Posted September 22, 2017 Author Report Posted September 22, 2017 the pcb located on the PCI slot in the PC have a blue switch called DS1 (switch 1 and 2), what is the right combination please? thank you
Minilab service Posted September 22, 2017 Report Posted September 22, 2017 ARCNET test software is on C Noritsukoki Self check. Before running it remove QSS main software shortcut from Start , Programs, Startup.
vincent Posted September 22, 2017 Author Report Posted September 22, 2017 (edited) here the result Communication is incorrect with all the connected ARCNET devices. Confirm the following items. - Selected(Model) -ARCNET cable connection to the PC - ............................................to the PCB with HUB -................cable connection - ............. cable for dirt - .............cable connector for dirt if the connection is still poor, possible cause for the problem is malfunction of the PCB on which arcnet chip is installed Edited September 22, 2017 by vincent
Minilab service Posted September 22, 2017 Report Posted September 22, 2017 Then no connection between computer board and Digital ICE control pcb. Remove scanner ARCNET cables from Digital ICE control pcb ( white color cable ) . Then connect it to computer pcb directly. Run ARCNET diagnostic software. You should have good connection with scanner ( scanner should be powered ) . If you will not have communication with scanner, then something wrong on computer. If have communication with scanner then problem with digital ICE control pcb. During test have select correct minilab model ( QSS32xx ) . Otherwise you have error. By the way - I wrote about normal QSS3202. If you have QSS3202SD minilab then connection is different.
vincent Posted September 23, 2017 Author Report Posted September 23, 2017 (edited) This morning, today i have a second error : 6901-2008 Processor D-Ice control PCB communication error error 6901-0008 appears first when i cancel it, 6901-2008 appears so i made a arcnet test again, i have (IMAGE PROCESSING PCB 7/7 errors, the other have not error) - ARCNETcable connection - ARCNET cable for dirt - ARCNET cable connector for dirt if the line state is still poor, possible cause for the problem is malfunction of the PCB on which arcnet chip is installed when you said Then connect it to computer pcb directly . you mean connect directly the scanner to PC? white cable and optical fiber ? then make another Arcnet test ? Edited September 23, 2017 by vincent
Minilab service Posted September 23, 2017 Report Posted September 23, 2017 (edited) Yes directly to computer, just optical cable. Do not touch LVDS cable. Remove scanner ARCNET cables from Digital ICE control pcb. Scanner ARCNET cable is white. Connect them directly to computer. Run ARCNET test. Should not be errors with scanner. This way you test computer board. If have communication with scanner then check all step by step. To computer connect original cables ( they go to Digital ICE control pcb ) . From this board disconnect all ARCNET cables ( except from computer ) . Run ARCNET diagnostic. This time should not have error with Digital ICE control pcb. Edited September 23, 2017 by Minilab service
vincent Posted September 25, 2017 Author Report Posted September 25, 2017 Hi, i connect the scanner directly to the computer. I have 2/7 errors. But my scanner is wrong, did not work for a while. So i don't know if it is computer board or not default Then i connect the others optical to the computer directly. I have errors for all. So it look like my digital control ICE pcb is wrong.
Minilab service Posted September 25, 2017 Report Posted September 25, 2017 What didn't work on your scanner? For this test working ARCNET is enough. This software doesn't test other scanner parts. Your answer was ''very informative'' . ''I have 2/7 errors'' says nothing. Do you mean your didn't have errors with PCI card and scanner and had communication errors with other boards? If yes - than it is normally. To your system connected only scanner, so all other boards should have errors. Conclusion - inside computer all works well. Scanner ARCNET communication is working also. Do not know what is ''connect the others optical to the computer directly'' . As I told before, if PCI card and scanner are OK then to computer connect ORIGINAL cables ( they go to Digital ICE control pcb ) . From Digital ICE control pcb DISCONNECT all ARCNET cables ( except cables from computer ) . Run ARCNET diagnostic. This time should not have error with Digital ICE control pcb and have errors with all other boards.
vincent Posted September 25, 2017 Author Report Posted September 25, 2017 (edited) I can not open noritsu soft, so i can not say the error from the negative scanner S2. I remember it was about: image processing error..... So i can not print negative for a moment, even can not do daily setup. sincerely sorry , i may be not understand you. I ask you again the procedure please. first: disconnect only the white optical fiber from PCB and connect directly to behind PC. run selfcheck second: i have 4 optical fiber from up to down . The first is the white one (scanner S2), the second is going to another PCB on the left, the third is going to computer and the fourth is going to printer. I disconnect all the optical fiber except cables from computer. run selfcheck is that correct ? Edited September 25, 2017 by vincent
Minilab service Posted September 25, 2017 Report Posted September 25, 2017 Do not know which cable - 1, 2 or some other. Cables can be connected to ANY port. Yes, from Digital ICE control pcb remove all ARCNET cables. Leave connected only ARCNET cables, which go to computer. Run ARCNET test. Look testing results. You should NOT have errors with Digital ICE control pcb. All other boards will have errors . It is normally, because they are disconnected. Only if have all that you can go forward.
davidlam Posted September 25, 2017 Report Posted September 25, 2017 Dear Vincent, how long have you been using your QSS-3202 ? Was it suddenly became "slow" one day? or you bought it and found it slow from day 1 ?
vincent Posted September 25, 2017 Author Report Posted September 25, 2017 hi, i've been using 3202 for more than 10 years. Prints become slow since last year and sometimes stop printing
Shady Posted April 19, 2019 Report Posted April 19, 2019 Hello, i have been using 3702 HD before i move it to another place, i tried to power it up again but the error (6901 Arcnet communications error) face me. what can i do for that error
Minilab service Posted April 19, 2019 Report Posted April 19, 2019 First what you should on this forum start new topic about ARCNET problem on QSS3702 HD . Press green button where see "Start new topic" . Only there can answer all your questions. This topic is about QSS3202.
Kassy Posted November 30, 2019 Report Posted November 30, 2019 Bonjour, Depuis une semaine ma machine noritsu qss3501 plus affiche un message d'erreur ARCNET Communication error, j ai essayer de changer le disk dur d'echange, changer la carte réseau intégré, et même revoir tout les câble du branchement rj45 tout rien, j'ai accuse la carte d'alimentation de 34vl , merci de votre aide Serge kassy kasias
Minilab service Posted November 30, 2019 Report Posted November 30, 2019 Do not understand your language. This is international forum, so write in English. You should respect other forum members, if want to get answers to your questions. Marian 1