JOEL VANHEMEL Posted December 6, 2017 Report Posted December 6, 2017 (edited) I am experiencing this error (07117-00001), has anyone had this problem? If so did replacing the pressure pump unit resolve the issue? Edited December 6, 2017 by JOEL VANHEMEL
Minilab service Posted December 6, 2017 Report Posted December 6, 2017 07117-00001 - the pressure sensor did not detect dark. Can be faulty pressure pump, pressure sensor, I/O PCB, or even main control pcb. First check sensor - open it and close and see changes on input check menu. If sensor OK then check output. Use multimeter to messure and run pump from output check. Look circuit diagram to see how all connected.
JOEL VANHEMEL Posted December 6, 2017 Author Report Posted December 6, 2017 thanks for the quick reply. I was limping along and then a couple inks stopped flowing. I think this points to pressure pump. Thoughts?
Minilab service Posted December 6, 2017 Report Posted December 6, 2017 I do not know. Can be two possibilities - pressure pump doesn't work, or it do not get signal from board. As I told before take multimeter, run pump from output check and look what you have there. Can measure on the board - will be more easy access than pump.