2901- Yellow splotches in gray areas


Recommended Posts

Hi Dave, yes, this is happening on all paper widths, sizes, and paper types.

5" Luster

6" Gloss, Luster

7" Luster

11" Luster

The funny thing is that the daily setup does not have this issue because the gray scale isn't fine enough (or light enough) to have this problem show. 

Here is our daily setup from this morning :

     daily setup results:  https://imgur.com/N6qMe5i

     Daily setup : https://imgur.com/XbsIPQe

The problem started 2 days ago.  I tried to replace the Image Processing Board, but same problem, so it's happening AFTER the Image Processing Board.  The MLVA lamp is a month old. - Haven't checked the dichroic filters or MLVA head (don't know how)

 

 

 

 

Link to post
Share on other sites

I'm wondering if the CD is a bit weak. I would expect a higher black reading. Something like 2.10 is ideal.

To me it looks like a bad LUT problem. So worth trying to load in your backup data. You can always make a backup of what you have now before doing that.

I personally would do an all data backup in service mode.

Link to post
Share on other sites

Question on backup data restore - any guidance on WHAT to restore ? Or restore ALL data ?

We've been doing service level all data backups whenever there's a "change" to the system - and not nightly / weekly backups - so the last backup I have is from December ! of 2017  :  :wacko:

Link to post
Share on other sites

90 seconds on the rep pump will not do much. Remember the tank size is about 20 litres (assuming it's not SM).

You would need to replace about 2 litres of CD to see a difference.

I would restore all data, it doesn't matter too much if it's that old, do a few daily setups and see if the problem has gone. It will prove whether the problem is data related or not.

Link to post
Share on other sites

Awesome !  Performed the full system all data restore, 2 daily setups to get it back in balance, print out the 2 test prints and no more yellow bits !!

I Thank You very much Dave S !

I just need to perform those backups a bit more frequently. 

Now to balance all the papers and do another service level all data backup !

:D:1310_thumbsup_tone1:

Link to post
Share on other sites

Ok, less 8 days later, the problem has returned.

Performed another data restore - and this DID FIX IT.

The question now, is why ?  Why did it happen again, so quickly ?

How can we prevent this issue going forward, or is this a precursor to "something else" ??

I cloned and installed a new 128GB SSD at the end of November 2017 -- so the hard drive is "good".

 

 

Link to post
Share on other sites

The first thing to check is the chemistry is spot on.
For example if the BF is weak it can struggle to control the light areas on the 18 step setup prints. Check the rep pumps are outputting correctly.

Next thing to check is the colorimeter unit, make sure the pressure plate is good and is pushing the print up towards the colorimeter head.

I wondering how you do a trim maintenance on the SSD drive when operating with Windows 2000?  I personally would be wary of using an SSD with Windows 2K.
The machine does a huge amount of temporary file writing when printing, so this could be causing problems with an SSD.
It may be worth going back to a standard SATA drive just to eliminate this as the cause of the problem.

 

Link to post
Share on other sites

Chemistry seems to be OK, no major ups or downs on the daily setups (sorry, we don't run control strips).  Chem usage is good, but I will double check the pump output amounts later today.

Colorimeter unit and position is good -- Only issue is the calibration plate is a bit on the old side. -- Last replaced back in Oct of 2009.

TRIM isn't/can't be enabled on Win2K (from what I could find out) but no un-healthy side effects from it so far.  Other people say that the system might be a little slower, but it's more than fast enough for us.  The usable space has not deviated at all in the past 8 months - that I can tell, but I've not really thought about it. -- I do have a good working HDD inside the machine, if the SSD ever fails, I just have to move the SATA/power cable from the SSD to the HDD, do a system restore and we're back up.

BTW, it's an 80GB SanDisk SSD, not 128gb.

 

Link to post
Share on other sites

Hi Dave S,

Replenishment pump output amount checks - looks mostly good. :unsure:  Maybe under replenishing a tiny little bit.

previously: CD = 60.0 / BF = 58.1 / STB = 111.0

NOW: CD = 58.3 / BF = 57.0 / STB = 113.3 - each run 4 times, measured and divided by 4

 

As far as the SSD TRIM function, there are quite a few others that have mentioned that they are running on older OS' with SSDs and not having issues after use for over 2 years -- and these were posts from 2015.  This article, from 2015, seems to say that even without TRIM, it might wear out your SSD a little faster, but nothing to really worry about.  https://arstechnica.com/gadgets/2015/04/ask-ars-my-ssd-does-garbage-collection-so-i-dont-need-trim-right/

The SanDisk I installed was bought in 2017, so should be good for at least another 10 years - unless it has a catastrophic failure at some point -- it was on sale when I bought it !  -- advertised as an 80 GB drive, 75 GB total with 64 GB free. -- lots of spare cells to use.

I can swap SSD to HDD if you think that's a plausible suspect in this data corruption.

Thanks !

 

Link to post
Share on other sites
  • 1 year later...

Hi Dave S !

Merry Christmas, happy new year and all the other cheery stuff that happy people say to other happy people.....

Finally got it nailed down, identified, replaced, and everything is BETTER than before.

Past month or so, our daily setups have been ALL over the place.  Made me think it wasn't saving the previous day's values so that it was starting from scratch EVERY day.  Would take 3-7 daily setups to get to "ok" values close to +/- 0.0.... the initial daily setups had correction values of up to +20, then -19, the +16, until it got back to 0.... if you add up all those numbers it's pretty much back at 0 where it started from, but it looks "normal", not white like the first print.  It started much lower, but the # of daily setups remained fairly constant at 3 to 5 times every day, just the initial correction values kept creeping up every day.  One day, we had to do an emulsion # change on the 6" matte, and it was bad.   After the 3rd fail, the top 3 or 4 lines were a bright yellow.  Didn't want to continue, so cancelled out, tried printing something, it came out mostly brown with a couple of black spots and looked like doo-doo.

So, started looking for answers, to "yellow prints", and found these:

 

Immediately changed out the Drive Voltage PCB, and the magenta cast we've had on our daily setups (that we've asked everybody how can we "fix" it) is gone !  As is the Yellow "splotches" in greys that we've had here.  Every thing seems to be back to normal (its been a week now).

Just thought you would like to know how we got rid of it.

Have a great holidays, enjoy your time off (if you get any!) and be well !

Big Dave

 

Link to post
Share on other sites

Hi Big Dave!

Thanks for posting your findings.

The drive voltage PCB is a common failure when you are getting big sudden colour shifts, however it tents to show up with the prints going red or brown. Turning the power off and back on again and the colour suddenly corrects are classic symptoms of this PCB being the cause of the problem.

I guess in early failure the colour shift results maybe are not so dramatic.

Merry Christmas, and a Happy new year to you too.

Link to post
Share on other sites

If on QSS29  color suddenly changes to brown - red, or similar color then drive voltage pcb is faulty. Many times had to repair boards with this fault.

If have red, brown, yellow dots, or small lines - usually one image correction pcb is faulty. If to switch boards ( two the same boards on QSS29 ) - often see error 6820-0004. The same image correction pcb used on QSS28, QSS29, QSS31 and QSS34 series minilabs ( not on all models ) .

Link to post
Share on other sites
  • 1 year later...

Hello gentlemen,

It seems it's back, but different, and I can't figure out how to fix it.   

Currently the daily setup is swinging wildly (within a certain range) from [Y+4.2, M+1.0, C+0, D+1.0] to [Y-4.2, M-1.2, C+0, D-1.2]

The setup print swings between blue and yellow/brown.

ex: past 5  days Daily setup values were : 

            Y        M       C        D

          -2.4    -0.4    0.0    -0.4

          +3.7    +0.4    0.0    +0.7

          -4.0    -0.6    0.0    -0.8

          +3.5    +0.5    0.0    +0.6

          -4.4  -0.8  0.0  -1.0  (today)

After the daily setup reads the daily setup print, it accepts the values no matter how large when it should fail and PRINT the daily setup print again.

Cyan is always +0.0.  performing emulsion # changes, per paper magazine, will also exhibit similar swings for each emulsion # change print which fails, but accepts the 3rd printout no matter its values, so I quit doing those.  

Instead, I adjust the balance memory settings in Paper Setup.  Not fun, easy or quick.

I got a new Drive Voltage PCB, from @davidlam, but replacing did not solve the problem.

 

What should be next ??

 

Thanks in Advance !

 

Link to post
Share on other sites

It will never ask you to daily setup again no matter how far out it is!

It will always accept the Emulsion test on the 3rd print, it figures if it’s not correct after the 3rd print it will not be able to get it any better.

It never gives a cyan correction.

 

Are the readings from the colorimeter plausible to the test print?

I’m wondering if the colorimeter is reading properly.

It seems like the readings are opposite from each other.

 

Does it ever get the print back to neutral?

Print a daily setup print but don't read it.  Shutdown the machine (Power off in close down check F functions) switch it off the breaker switch for 5 mins, then switch it back on and do another daily setup, does the print match the first daily setup?

Link to post
Share on other sites
4 hours ago, Dave S said:

Are the readings from the colorimeter plausible to the test print?  --  I am guessing "YES" -- when the test print is "blue-ish"  the correction values will be Yellow (+) POSITIVE (ie: Y +3.8); when the test print is "yellow/ brown" the correction values are NEGATIVE (ie: Y -4.0)

It seems like the readings are opposite from each other.  -- yes, they "Flip-Flop'

Does it ever get the print back to neutral? -- NO.  After daily setup, I print a DIMA calibration print (grey scale and couple of photos with grey border)  and the calibration print will be opposite of the daily setup test print. 

ex: daily test print is blue-ish, and the  correction values are: +3.9 / +1.0 /  0.0 / +1.2.   Printing the calibration print, it would come out brownish, ie : too much yellow.  So adjust the  balance memory settings in Paper Setup with -3.9 / -1.0 / 0.0 / -1.2, and reprint the calibration print, and it comes out closer to Normal, and then just repeat, print/tweak/print/tweak a few more times until they're somewhat close to neutral.  Then do it all over again the next day. 

If the weekly setup is performed, all the values get mixed up again, and you start from a blank slate and it's even harder to get to neutral.

Print a daily setup print but don't read it.  Shutdown the machine (Power off in close down check F functions) switch it off the breaker switch for 5 mins, then switch it back on and do another daily setup, does the print match the first daily setup?  --  will let you know tomorrow.

 

Link to post
Share on other sites

@Big Dave

# Z021440-01 Calibration plate unit A (for QSS-2901 & 3001 ..) had been discontinued for a long time and SKY PRICE before its discontinuance... hard to find new one

    I can help to find new SET   if   EXPERTs  here say it can help the color stability

# what type is your QSS-2901 Colorimeter model ?   CS-CM1000,  CM1000M,  CM1000N, Q, M, R or S  type.   there is a label on the hand-hold colorimeter telling that

   be very very careful when you open it Up for cleaning  ... as the inside color filterssss  will fall out easily and you may not tell which is which position .. my pain before

# Could it be a "Filter Wheel" issue for the color "Flip-Flop"  ???   I doubt 

Link to post
Share on other sites

Thanks @davidlam, but I do have a new one.... (I think last one Noritsu America had sold to friend in SanDiego, then sold to me "semi-used")...

Bought 2, or 3 years ago ?  dirt cheap compared to Noritsu direct.

 I've never opened the colorimeter.

It could be the filter wheel but I don't know if it would create these issues.  Last time it was generating lines on the prints, but that was because it was slowly cracking and was partially damaged.   Mine's an old, used replacement from a different machine.

Link to post
Share on other sites

Hi @Dave S,  Yes, 

20 hours ago, Dave S said:

Print a daily setup print but don't read it.  Shutdown the machine (Power off in close down check F functions) switch it off the breaker switch for 5 mins, then switch it back on and do another daily setup, does the print match the first daily setup?

Not 100% but close enough.

for mlh.jpg

Link to post
Share on other sites

Oh, also, the Daily Setup test print is using 152(1) 6" Glossy  BUT  we print (mainly) using 152(2) LUSTER   and 279(2) LUSTER and these are the 2 paper magazines we manually "bring to neutral" (or a close approximation of it) we stopped using Glossy for customer prints.

Doing Emulsion # changes on either of the Luster papers is a losing battle -- 3 test prints and accepts the 3rd one no matter how bad it is.

 

Link to post
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...