Jack Johnson Posted December 12, 2022 Report Share Posted December 12, 2022 We have two Noritsu HS-1800 scanners with separate computers and EZ-Controllers. We had both scanners adjustet and calibrated by a Noritsu technician a couple of months ago, but we have now found out that both scanners stretch the scans by aprox. 5% when we are scanning 120. Any thoughts and could it be anything else than a wrong calibration? Quote Link to comment Share on other sites More sharing options...
davidlam Posted December 12, 2022 Report Share Posted December 12, 2022 2 different 120AFC ? IF two, try swop / swap the 120AFC to see if problem goes to another HS-1800 .... This can isolate AFC roller possibilities could be one 120AFC still with genuine rollers while another 120AFC with China-made rollers Quote Link to comment Share on other sites More sharing options...
Jack Johnson Posted December 12, 2022 Author Report Share Posted December 12, 2022 Two different 120AFC - one for each scanner. (The scanners are bought from two different suppliers) We have tried to use "120AFC A" on "Scanner A", "120AFC B" on "Scanner A" and vice versa. The stretch is exactly the same. Quote Link to comment Share on other sites More sharing options...
Dave S Posted December 12, 2022 Report Share Posted December 12, 2022 It is normally caused by non genuine rollers that end up slipping on the shaft. Quote Link to comment Share on other sites More sharing options...