SP3000 frame position error in auto feeder


CatLABS

Recommended Posts

Recently, my SP3000 has been missing frame position, progressively as the roll advances through the autofeeder.

The first 6-10 frames are OK, then they start to randomly shift, some more than others, sometimes so much that part of the previous frame is in the scan. 

This happens despite manually going over each frame in the preview and adjusting its position.

 

How can i correct this?

 

Thanks

image0 (1).jpeg

Edited by CatLABS
Link to comment
Share on other sites

It is possible to adjust only position of all frames. When advance is unstable you have some film carrier mechanical fault and need repair it. Film is advanced instable. At the beginning small part of film is inside film carrier and it is more easy to make film advance. When bigger part of film is inside it is more difficult make film advance. When something inside film carrier doesn't work properely film is supplied instable. Might be something disturb to move film on second film carrier part, but can be that have film advance instability everywhere and instability accure when inside film carrier is bigger part of film.

Link to comment
Share on other sites

On 3/3/2023 at 4:16 AM, Minilab service said:

It is possible to adjust only position of all frames. When advance is unstable you have some film carrier mechanical fault and need repair it. Film is advanced instable. At the beginning small part of film is inside film carrier and it is more easy to make film advance. When bigger part of film is inside it is more difficult make film advance. When something inside film carrier doesn't work properely film is supplied instable. Might be something disturb to move film on second film carrier part, but can be that have film advance instability everywhere and instability accure when inside film carrier is bigger part of film.

Yes, this is true, the later part of the film position gets progressively worse as the film advances through the scanner.

Can you hel me understand where the problem is and how to fix it?

Link to comment
Share on other sites

Need to find what cause instability and repair film carrier. Here is mechanical problem. Impossible to gues what exactly cause instability when do not have film  carrier myself.

Using software settings is possible when frame position difference is stable ( for example 1st frame 1 mm, 2nd frame 2mm, 3rd frame 3 mm etc ) . 

Link to comment
Share on other sites

5 hours ago, Minilab service said:

Need to find what cause instability and repair film carrier. Here is mechanical problem. Impossible to gues what exactly cause instability when do not have film  carrier myself.

Using software settings is possible when frame position difference is stable ( for example 1st frame 1 mm, 2nd frame 2mm, 3rd frame 3 mm etc ) . 

Where can i find the software setting for the frame offest?

Link to comment
Share on other sites

Probably you are talking about offset ( not about offest ) . This setting is not related with film position. NC100AY focus offset adjustment - menu 0441.

Frame position correction
0400 - NC100AY fixed feeding setup
0420 - mask position adjustment
0440 - NC100AY sensor calibration
0451 - NC100AY machine data setup
Frame position can be corrected only if film advance is stable and always film stopps in the same position.

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...