tshicode Posted April 21 Report Share Posted April 21 Good morning, I have a problem with my noritsu 3311 machine, it displays a message "laser temperature are being adjusted (r)" Can you help me Quote Link to comment Share on other sites More sharing options...
Minilab service Posted April 21 Report Share Posted April 21 Hello. If you see this message not very long time then it is normally. Wait about 10 minutes. Might be this message disappear. If not then go Menu → Extension → Maintenance → Input Check → Printer → Engine Section . There look Laser Light Source Status, Laser Judgment Signal 1 , Laser Judgment Signal 2 and Laser Ready Status . Look both laser heads ( G and B ) . When checking status laser should be already preheated. Not all of them are when laser is B1 type, but I think you have A1 or Ff type laser. Laser type can see on software version check menu ( line where is laser version ) . Probably one your laser head is faulty and you need to change it. tshicode 1 Quote Link to comment Share on other sites More sharing options...
Dave S Posted April 21 Report Share Posted April 21 In my experience this problem usually comes from a bad connection for the Red laser thermosensor. Go to Menu → Extension → Maintenance → Input Check → Printer → Engine Section and Check what the R laser temperature is, and what the laser unit temperature is displaying. tshicode 1 Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 22 Author Report Share Posted April 22 Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 22 Author Report Share Posted April 22 19 hours ago, Minilab service said: Bonjour. Si vous voyez ce message pas très longtemps, c'est normalement. Attendez environ 10 minutes. Peut-être que ce message disparaîtra. Si ce n'est pas le cas, allez dans Menu → Extension → Maintenance → Vérification d'entrée → Imprimante → Section Moteur. Vous y trouverez l'état de la source lumineuse laser, le signal d'évaluation laser 1 , le signal d'évaluation laser 2 et l'état prêt au laser . Regardez les deux têtes laser (G et B). Lors de la vérification de l'état, le laser doit déjà être préchauffé. Tous ne le sont pas lorsque le laser est de type B1, mais je pense que vous avez un laser de type A1 ou Ff. Le type de laser peut être vu dans le menu de vérification de la version du logiciel (ligne où est la version du laser). Probablement une de vos têtes laser est défectueuse et vous devez la changer. Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 22 Author Report Share Posted April 22 19 hours ago, Dave S said: D'après mon expérience, ce problème provient généralement d'une mauvaise connexion pour le thermocapteur laser rouge. Allez dans Menu → Extension → Maintenance → Vérification d'entrée → Imprimante → Section Moteur et vérifiez quelle est la température du laser R et quelle est la température de l'unité laser affichée. Quote Link to comment Share on other sites More sharing options...
Minilab service Posted April 22 Report Share Posted April 22 Red laser temperature is too low. Check voltage on LASER I/O PCB connector J/P1540 ( d.c. between pin 1 and pin 2 ) . When R laser warming on pin 1 should be "-" and on pin 2 should be "+" . Voltage will be up to 5V ( usually it is lower ) . Check Red laser head termosensor. Between ground and LASER I/O PCB connector J/P1532 pin 7 should be 1V ( reference voltage ) . Here connected sensor input. Between ground and J/P1532 pin 8 voltage depends on R laser head temoerature. Here connected sensor output. Try turn off minilab and reconnect connectors J/P1532, J/P1536 and J/P1540, J/P1541 on LASER I/O PCB and J/P1511 on laser control pcb. Check LASER I/O PCB power suppliers : between ground and J/P1535 pin 1 and 3 should be 5V between ground and J/P1541 pin 1 , 2 and 3 should be 24V between ground and J/P1550 pin 1 should be 12V ( positive voltage ) between ground and J/P1550 pin 4 should be -12V ( negative voltage ) Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 22 Author Report Share Posted April 22 16 minutes ago, Minilab service said: La température du laser rouge est trop basse. Vérifiez la tension sur le connecteur PCB LASER I/O J/P1540 (cc entre la broche 1 et la broche 2). Lorsque le réchauffement du laser R sur la broche 1 doit être "-" et sur la broche 2 doit être "+". La tension sera jusqu'à 5V (habituellement elle est inférieure). Vérifiez le capteur thermique de la tête laser rouge. Entre la masse et le connecteur PCB I/O LASER J/P1532, la broche 7 doit être de 1 V (tension de référence). Entrée de capteur connectée ici. Entre la masse et la tension de la broche 8 de J/P1532 dépend de la température de la tête laser R. Voici la sortie du capteur connecté. Essayez d'éteindre le minilab et de reconnecter les connecteurs J/P1532, J/P1536 et J/P1540, J/P1541 sur la carte d'E/S LASER et J/P1511 sur la carte de contrôle laser. Vérifiez les sources d'alimentation du PCB I/O LASER : entre la terre et J/P1535 les broches 1 et 3 doivent être de 5 V entre la terre et la broche 1 de J/P1541, 2 et 3 doivent être de 24 V entre la terre et la broche 1 de J/P1550 doivent être de 12 V (positif tension) entre la masse et la broche 4 de J/P1550 doit être de -12 V (tension négative) merci beaucoup vous m'avez vraiment aider Quote Link to comment Share on other sites More sharing options...
Minilab service Posted April 22 Report Share Posted April 22 Do not understand what you wrote here. Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 22 Author Report Share Posted April 22 il y a 54 minutes, Minilab service a dit : Je ne comprends pas ce que vous avez écrit ici. merci de m'avoir aidé Thanks for helping me Quote Link to comment Share on other sites More sharing options...
Dave S Posted April 26 Report Share Posted April 26 Did you solve the problem? Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 26 Author Report Share Posted April 26 Le 21/04/2023 à 15h32, le service Minilab a déclaré : Bonjour. Si vous voyez ce message pas très longtemps, c'est normalement. Attendez environ 10 minutes. Peut-être que ce message disparaîtra. Si ce n'est pas le cas, allez dans Menu → Extension → Maintenance → Vérification d'entrée → Imprimante → Section Moteur. Vous y trouverez l'état de la source lumineuse laser, le signal d'évaluation laser 1 , le signal d'évaluation laser 2 et l'état prêt au laser . Regardez les deux têtes laser (G et B). Lors de la vérification de l'état, le laser doit déjà être préchauffé. Tous ne le sont pas lorsque le laser est de type B1, mais je pense que vous avez un laser de type A1 ou Ff. Le type de laser peut être vu dans le menu de vérification de la version du logiciel (ligne où est la version du laser). Probablement une de vos têtes laser est défectueuse et vous devez la changer. Le 21/04/2023 à 16:10, Dave S a dit : D'après mon expérience, ce problème provient généralement d'une mauvaise connexion pour le thermocapteur laser rouge. Allez dans Menu → Extension → Maintenance → Vérification d'entrée → Imprimante → Section Moteur et connaître quelle est la température du laser R et quelle est la température de l'unité laser affichée. Le 22/04/2023 à 11h25, le service Minilab a déclaré : La température du laser rouge est trop basse. Vérifiez la tension sur le connecteur PCB LASER I/OJ/P1540 (cc entre la broche 1 et la broche 2). Lorsque le nettoyage du laser R sur la broche 1 doit être "-" et sur la broche 2 doit être "+". La tension sera jusqu'à 5V (habituellement elle est inférieure). Vérifiez le capteur thermique de la tête laser rouge. Entre la masse et le connecteur PCB I/O LASER J/P1532, la broche 7 doit être de 1 V (tension de référence). Entrée de capteur connectée ici. Entre la masse et la tension de la broche 8 de J/P1532 dépendent de la température de la tête laser R. Voici la sortie du capteur connecté. essayez d'éteindre le minilab et de reconnecter les connecteurs J/P1532, J/P1536 et J/P1540, J/P1541 sur la carte d'E/S LASER et J/P1511 sur la carte de contrôle laser. Vérifiez les sources d'alimentation du PCB I/O LASER : entre la terre et J/P1535 les broches 1 et 3 doivent être de 5 V entre la terre et la broche 1 de J/P1541, 2 et 3 doivent être de 24 V entre la terre et la broche 1 de J/P1550 doivent être de 12 V (tension positive) entre la masse et la broche 4 de J/P1550 doivent être de -12 V (tension négative) non pas encore Quote Link to comment Share on other sites More sharing options...
tshicode Posted April 26 Author Report Share Posted April 26 non pas encore Quote Link to comment Share on other sites More sharing options...
davidlam Posted April 26 Report Share Posted April 26 English please. . Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 31 Report Share Posted May 31 Hi Try to check the AOM and Driver as well Thanks Quote Link to comment Share on other sites More sharing options...
Minilab service Posted May 31 Report Share Posted May 31 6 hours ago, Jhun said: Hi Try to check the AOM and Driver as well Thanks Do not write nonsense. AOM driver is not related with R laser head temperature. Even without AOM driver R laser temperature should be correct. Here it is lower, than should be. When one month, or longer period passed probably problem was solved. Do not know, why you wrote all your messages. Quote Link to comment Share on other sites More sharing options...