Jhun Posted May 8 Report Share Posted May 8 (edited) Hi! Does anyone experienced this kind of error? I opened already the Colorimeter unit and cleaned the advanced roller but still error appeared. for those who experienced this kind of error kindly comment/advice are highly appreciated.Thanks Edited May 8 by Jhun lack info Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 8 Report Share Posted May 8 What does the test print look like? Is the first step white and last step deep black? Quote Link to comment Share on other sites More sharing options...
PHOTOLABPART.COM Posted May 11 Report Share Posted May 11 RESPECTED SIR PLEASE SHOW LOADING PAPER BLACK AND WHITE IF CHEMISTRY AND PAPER IS OK CHECK THEIR CALIBRATION PLATE AND COLORIMETER UNIT THANKS Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 13 Author Report Share Posted May 13 On 5/9/2023 at 12:47 AM, Dave S said: What does the test print look like? Is the first step white and last step deep black? hi sir, it is normal YES sir it is white 1st step and BLACK on last Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 13 Author Report Share Posted May 13 On 5/11/2023 at 8:12 AM, PHOTOLABPART.COM said: RESPECTED SIR PLEASE SHOW LOADING PAPER BLACK AND WHITE IF CHEMISTRY AND PAPER IS OK CHECK THEIR CALIBRATION PLATE AND COLORIMETER UNIT THANKS yes sir I tried to feed a paper it is black is black and white is white Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 14 Report Share Posted May 14 20 hours ago, Jhun said: hi sir, it is normal YES sir it is white 1st step and BLACK on last Can you scan the setup print with a flatbed scanner, and upload the photo on here or use a file sharing site (wetransfer etc) and share the link. Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 15 Author Report Share Posted May 15 22 hours ago, Dave S said: Can you scan the setup print with a flatbed scanner, and upload the photo on here or use a file sharing site (wetransfer etc) and share the link. ok sir Thanks Quote Link to comment Share on other sites More sharing options...
PHOTOLABPART.COM Posted May 15 Report Share Posted May 15 RESPECTED SIR SCAN COPY LOAD PAPER ALSO FOR LOOK BLACK AND WHITE THANKS Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 18 Author Report Share Posted May 18 I tried to change the chemical but still Quote Link to comment Share on other sites More sharing options...
PHOTOLABPART.COM Posted May 19 Report Share Posted May 19 PLEASE SHOW BLACK AND WHITE OR GAMMA PRINT Quote Link to comment Share on other sites More sharing options...
Minilab service Posted May 19 Report Share Posted May 19 Try to measure your test on other minilab. Can mesure on any QSS32, QSS33, QSS34 minilab. Make screenshot with results. Of cource there do not enter mesurement results ( make screensot and cancel ) . Then you know where is the problem - test, or your minilab colorimeter. Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 21 Report Share Posted May 21 On 5/18/2023 at 6:24 PM, Jhun said: I tried to change the chemical but still Without seeing what your test print looks like, I can't offer any further advice. Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 22 Author Report Share Posted May 22 Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 22 Author Report Share Posted May 22 hi sir here is the print out and I tried to read to another machine 3300 same message Thanks Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 22 Report Share Posted May 22 When you changed the chemistry did you add starter to the CD? It looks like the CD mix is too strong. Possibly the BF may also be bad. Quote Link to comment Share on other sites More sharing options...
oliphaz Posted May 22 Report Share Posted May 22 Yeah this error isn't necessarily an issue with the colorimeter, the plots could just be outside of correctable bounds. Have you run control strips? I might try running a new paper registration. Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 23 Author Report Share Posted May 23 23 hours ago, Dave S said: When you changed the chemistry did you add starter to the CD? It looks like the CD mix is too strong. Possibly the BF may also be bad. Hi sir last week and I tried to mix again and perform initial set up but still error appear Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 23 Author Report Share Posted May 23 23 hours ago, oliphaz said: Yeah this error isn't necessarily an issue with the colorimeter, the plots could just be outside of correctable bounds. Have you run control strips? I might try running a new paper registration. i perform already initial set up sir Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 23 Report Share Posted May 23 What chemistry are you using? Did you follow the instructions to make tank solution, or did you mix it the same as for replenisher? Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 23 Report Share Posted May 23 Have you replaced the BF chemical? Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 23 Author Report Share Posted May 23 12 minutes ago, Dave S said: Have you replaced the BF chemical? yes sir just today Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 23 Author Report Share Posted May 23 14 minutes ago, Dave S said: What chemistry are you using? Did you follow the instructions to make tank solution, or did you mix it the same as for replenisher? tank solution instruction sir and adding a starter Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 23 Author Report Share Posted May 23 I cannot upload the photo instructions in the box Quote Link to comment Share on other sites More sharing options...
Dave S Posted May 23 Report Share Posted May 23 (edited) The print doesn't look right, to me it looks like a chemistry issue. Maybe the chemistry is old and out of date, or is for the wrong process, who knows. The only way to check the chemistry is properly correct is to processes and read a paper control strip. https://www.kodak.com/global/plugins/acrobat/en/service/Zmanuals/z130_08.pdf Edited May 23 by Dave S Quote Link to comment Share on other sites More sharing options...
Jhun Posted May 25 Author Report Share Posted May 25 On 5/22/2023 at 9:57 PM, oliphaz said: Yeah this error isn't necessarily an issue with the colorimeter, the plots could just be outside of correctable bounds. Have you run control strips? I might try running a new paper registration. I already executed sir light intensity it was read 2nd print cannot Quote Link to comment Share on other sites More sharing options...