Hadi Posted July 12 Report Share Posted July 12 I need help about this message" The processing solution safety thermostat has activated" Quote Link to comment Share on other sites More sharing options...
Dave S Posted July 12 Report Share Posted July 12 Which machine model? It means the processing solution has got too hot in the subtank. Check the chemical filters are not blocked and the circulation is good. Quote Link to comment Share on other sites More sharing options...
Hadi Posted July 12 Author Report Share Posted July 12 It's the qss3300 ok I'll check the filters and the chemistry circulation. Thanks Quote Link to comment Share on other sites More sharing options...
Hadi Posted July 18 Author Report Share Posted July 18 (edited) Thank you very much Dave S for your reaction. The problem is solved. I simply left the BF open all night and the next day the error message did not come. And I left it open all day and I also lowered its temperature to 35C° and that's it. Edited July 18 by Hadi Quote Link to comment Share on other sites More sharing options...