Reynaldo Posted February 20 Report Posted February 20 Hello, good day. I hope everyone is doing well. If possible, could you help me with a problem I'm having with my Fuji 7700 unit? When I turn it on, it displays six errors, all related to the circulation pumps. For example: No. 05511-00000 No. 05513-00000 No. 05515-00000 No. 05516-00000 No. 05517-00000 No. 05518-00000 The rotation detector built into the P1 circulation pump detects that the pump has stopped rotating. Work performed, power supply replacement: 24V/9A with 5V/5A ZWD225PAF 24V/6.3A ZWS15OBAF-24 J391466 board J391468 board And the same error persists. I'm a bit lost. If possible, could you share your experience? Regards and thanks
Dave S Posted February 20 Report Posted February 20 Are the circulation pumps physically running or not? Reynaldo 1
Reynaldo Posted February 20 Author Report Posted February 20 Good idea, I'll check if the pumps are working. Thanks in advance.
Reynaldo Posted February 20 Author Report Posted February 20 Hola buen dia , mucha gracias voy a remplazar esa tarjeta en contacto
Reynaldo Posted February 20 Author Report Posted February 20 (edited) Hace 1 hora, kodak_service dijo: Intente reemplazar la placa J391467. Hola Buen día, cambia tarjeta y me marco misma falla en contacto saludos. Edited February 20 by Reynaldo
kodak_service Posted February 20 Report Posted February 20 Check the fuse F13 on J391466 board. Reynaldo 1
Reynaldo Posted February 20 Author Report Posted February 20 12 minutes ago, kodak_service said: Check the fuse F13 on J391466 board. Hola buen dia muchas gracias , cambie la tarjeta y el fusible se encuentra bien en contacto.
kodak_service Posted February 20 Report Posted February 20 I didn't understand you. Is the problem solved or not? Reynaldo 1
Reynaldo Posted February 20 Author Report Posted February 20 Hola una disculpa no me explique bien, cambie esa tarjeta por una nueva y se checo el fusible y esta correcto , pero tengo la misma falla el problema no esta resuelto .muchas gracias en contacto.
kodak_service Posted February 20 Report Posted February 20 (edited) Measure the voltage at the TP1 test point on the board J391466 or check the state of the LED14. Edited February 20 by kodak_service Reynaldo 1
Reynaldo Posted February 20 Author Report Posted February 20 Ok sip, voy a checar esos puntos , gracias en contacto.
Reynaldo Posted February 20 Author Report Posted February 20 2 hours ago, kodak_service said: Mida el voltaje en el punto de prueba TP1 en la placa J391466 o verifique el estado del LED14. Hello, I have a voltage of 23.8v and LED14 lights up green and fuse 13 is in good condition en contacto Gracias.
Reynaldo Posted February 20 Author Report Posted February 20 10 hours ago, Dave S said: ¿Las bombas de circulación están funcionando físicamente o no? Hello, the pumps are not physically circulating, as if there is something preventing them from activating.
Dave S Posted February 20 Report Posted February 20 (edited) . Edited February 20 by Dave S Reynaldo 1
kodak_service Posted February 20 Report Posted February 20 12 minutes ago, Reynaldo said: Hello, I have a voltage of 23.8v and LED14 lights up green and fuse 13 is in good condition en contacto Gracias. Measure the voltage at pins 1 and 4 of connector 651 on the J391466 board (relative to ground). Reynaldo 1
Dave S Posted February 20 Report Posted February 20 Check the resistance of the solution level float sensors between Pins 3 & 4 for the CD, BF and STB float switches. They should all be close to 0 ohms. The plug numbers are J/P784, 785, & 789 look for the black, blue and red wires from the float switches. Reynaldo 1
Dave S Posted February 20 Report Posted February 20 41 minutes ago, Reynaldo said: Hello, the pumps are not physically circulating, as if there is something preventing them from activating. The pumps are supplied voltage via 24V-1 which becomes 24V-1-1 and 24V-1-2. The relay X6 switches the 24V-1 power. I'm wondering if this relay is not being turned on. Reynaldo 1
Reynaldo Posted February 20 Author Report Posted February 20 1 hour ago, Dave S said: Verifique la resistencia de los sensores de nivel de solución entre los pines 3 y 4 de los interruptores de flotador CD, BF y STB. Todos deben estar cerca de 0 ohmios. Los números de enchufe son J/P784, 785 y 789. Busque los cables negro, azul y rojo de los interruptores de flotador. Gracias voy a checar en contacto
Reynaldo Posted February 20 Author Report Posted February 20 I have a question. When I turn on the main switch, the unit starts up, the dryer unit turns on, and then the unit resets to its home state. However, there's no water movement in the unit where the cylinder filters are located. After two minutes, an alarm is triggered, and errors appear for all the water pumps. But from pump P1, there's a noise as if something is stuck or obstructed. Could this be causing all the water pumps to alarm? Thanks.
Dave S Posted February 20 Report Posted February 20 25 minutes ago, Reynaldo said: I have a question. When I turn on the main switch, the unit starts up, the dryer unit turns on, and then the unit resets to its home state. However, there's no water movement in the unit where the cylinder filters are located. After two minutes, an alarm is triggered, and errors appear for all the water pumps. But from pump P1, there's a noise as if something is stuck or obstructed. Could this be causing all the water pumps to alarm? Thanks. Try unplugging the pump P1 and see if the error for the other pumps dissappears or not. Reynaldo 1
Reynaldo Posted February 27 Author Report Posted February 27 (edited) Hello, good day. I want to thank everyone for taking the time to share their knowledge. The problem has been resolved. The issue was a damaged processing solution float switch. It was replaced, and the problem no longer occurs. I think I panicked and went straight for the circuit boards. You're right, you should check the components first before replacing any boards. Thank you very much, and best of luck. Edited February 27 by Reynaldo Dave S 1
Dave S Posted February 27 Report Posted February 27 Yes there is a tendency for people to jump to the most complicated thing first, diagnosis is always better than just throwing parts at something and hoping for the best! Thanks for coming back and posting what the solution was to the problem. It's always nice to hear what the final solution was. Reynaldo 1