trevvvor Posted October 25 Report Posted October 25 Hello there. I have HS1800 that recently had a light source replaced maybe 3-4 months ago. I am not all of a sudden getting very dramatic red tones and my highlights are being blown out much quicker than on my other machines. I am attaching a sample, and images of Sensitivity Chart with/without shading on in hopes someone can confirm what might be going on here. Thank you!
trevvvor Posted October 25 Author Report Posted October 25 Ok, update: I have been running some EZ Controller adjustments to see what I can do, and I believe a Scanner Sensitivity adjustment and LED adjustment solved this color/highlight problem. Is this possible, or am I about to run into the same problem?
oliphaz Posted October 26 Report Posted October 26 The LED source looks good here. An increase in the gain values is what to look out for.
trevvvor Posted October 27 Author Report Posted October 27 On 10/26/2026 at 12:32 PM, oliphaz said: The LED source looks good here. An increase in the gain values is what to look out for. Thank you - I believe thats all it took to get it back running properly.
Dave S Posted October 27 Report Posted October 27 15 minutes ago, trevvvor said: Thank you - I believe thats all it took to get it back running properly. Any values you change in the sensitivity check mode do not get stored, they are just for testing purposes only. When you do the scanner calibration it will adjust these values as part of the calibration if it needs to. Most likely it just did it's auto scanner calibration and it corrected the issue. Doing the scanner calibration normally sorts out strange problems like this.
trevvvor Posted October 28 Author Report Posted October 28 30 minutes ago, Dave S said: Any values you change in the sensitivity check mode do not get stored, they are just for testing purposes only. When you do the scanner calibration it will adjust these values as part of the calibration if it needs to. Most likely it just did it's auto scanner calibration and it corrected the issue. Doing the scanner calibration normally sorts out strange problems like this. OK interesting - I believe I ran multiple Scanner Calibrations prior to running the sensitivity adjustment and the issue still remained. Either way I have ran a couple hundred rolls through since then and the issue has not come back